Calculating the mass of 3 ml 0.5N methanolic NaOH.
Posted: Tue May 10, 2011 4:18 pm
by Sandra7
I prepared 0.5 N methanolic NaOH weighing .2g of NaOH into 10 ml of MeOH. It took NaOH quite some times to dissolve.
I used 3 ml of this methanolic solution to mix with 0.08 g (80 mg) o extra virgin olive oil to perform saponification before derivatizing the mixture (olive oil + NaOH) with 12% (1.5M) BF3.
How would I go about calculating the mass of 3 ml methanolic NaOH used in saponification since I could see that the .5N mathanolic NaOH was a bit saturated? I took 3 ml of methanolic NaOH, making sure that the solid pieces of NaOH were left behind.
MW of NaOH = 39.99 g/mol
Density of NaOH = 2.13 g/ml
Re: Calculating the mass of 3 ml 0.5N methanolic NaOH.
Posted: Tue May 10, 2011 5:34 pm
by Consumer Products Guy
We also make up 0.5 M NaOH in methanol, it does take a while. I think the undissolved stuff is mostly sodium carbonate though. We compensate for the NaOH being 85% pure.
As to the density of the solution, you know the grams, can't you just dissolve up same in a stoppered graduated cylinder and measure the final volume?
Posted: Wed May 11, 2011 7:00 am
by Peter Apps
KISS - take another 3 ml from the approx 7 ml that you have left over, and weigh it.
What is this "N" ?
Posted: Wed May 11, 2011 9:02 am
by DSP007
Hello, well take a graduated cylinder at 10 ml and make three to 10 ml of 0.5 M solution - dissolve 0.2 g in 9 ml of methanol and methanol to bring the volume to 10 ml. Enough just on three parallel samples. Accuracy of preparation of this solution does not play a special role is reactive to saponification of fat. Plus-minus kilometer.
Posted: Wed May 11, 2011 9:08 am
by DSP007
What is this "N" ?
Hi
N- Normality of solution. Old termine from titrometry. C(mol/l) x Ionic activity (dissotiation factor) - for NaOH [0H-] in water -0.999 , for H2SO4 [H+ 2 const] in water 1.999(3).
From this quest N=M- precison of solution "plus- minus kilometer"
Posted: Wed May 11, 2011 10:21 am
by HW Mueller
Sandra7, you would have to titrate your NaOH solution to find out how much NaOH is in it. But what for (already questioned above)? Besides, weighing out 0.2 g of NaOH is not so hot . . . .
Why do you react with NaOH then BF3, transesterification with BF3/MeOH might do the trick for you.
Posted: Wed May 11, 2011 12:55 pm
by DSP007
Good afternoon, HW Mueller.
1) Yes you can mix a solution in a bucket and spade for this purpose. But why then do with it so much? So IMHO 0.2 g of sodium hydroxide
is just what the doctor ordered a pair of crumb or a Scale (granule).
2)Its Sandra methodology execute
Step right, step left - an attempt to escape, jumping in place provocation. (c) Although I myself do not like BF3 . IMHO ethyl esters is much nicer to work.
Posted: Sat May 14, 2011 12:23 am
Actually, HW, that would be what we do also. We've found that fats and oils (unlike fatty acid or soap samples) do not mix readily with methanol-BF3 or methanol-H2SO4, the triglyceride just sits on the bottom, not good contact with the methanol portion. So we saponify for 15 minutes, then acidify and heat a little to make methyl esters, have been doing this for three decades.
We haven't the need to try sodium methoxide or any fancy one-step reagents for this.
As to ethyl esters, one must use ethanol that contains no methanol or IPA denaturants, or you'll really get a mess. Fatty acid ethyl esters are typically used as retention index markers in the fragrance industry.
Posted: Sat May 14, 2011 9:12 am
by HW Mueller
CPG, do you have to work with such horrendously high concentrations for some reason?
Posted: Mon May 16, 2011 4:25 pm
No, we typically have lots of sample, take larger sample sizes to ensure homogeneity. When we do fats and oils, we use a 1 gram sample. With soaps or fatty acids, maybe 3 grams. We've done this for 35 years.
Posted: Mon May 16, 2011 7:08 pm
by chromatographer1
You are correct when you say the oils do not mix well with BF3/methanol.
I have transesterified oils and fats by adding DCM or Chloroform to the BF3/Methanol and heated the mixture. Then I only had to add my salt water to extract my esters.
Never had a problem. Are there problems with doing this? I would love to learm if there is a problem.
best wishes,