Separation Problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Dear All,

I am trying to separate polar isomeric compounds using LC-MS/MS;

I prepared a standard mixture of polar isomeric compounds ( configuration chemical group is -OH) of with known concentration in MeOH and optimized a method development

The compounds m/Z values after dansyl chloride derivatization are;

504/171
506/171
522/156
530/171 (Internal Standard)
536/171 and 534/171
753/170 and 755/521
520/171

The method is:
SP: C18; 25 cmx4.6 mm, 5um BECKMANN COULTER
MP: A: 95%Water/5% ACN with 0.1 % FA and

B is 95% Acetonitrile/ 5% Water with 0.1% FA

Flow rate: 800 uL/min

Gradient Programme is:

Time %B

At 0 min 0
5 min 40
8 min 50
12 min 55
19 min 65
25 min 100
31 min 0
35 min 0


PROBLEM IS:

All the compounds are eluting after 100% B only i.e., starting at 24 min, and with almost same RTs (Very little difference)


Any Idea how can get a good separation method with a shorten the analysis time given the current information.


-------------------
Raj
Hi
Isomers or stereoisomers? 8)

For stereoisomers separation- selection C18 column is a bad variant. Need use stereisomer phase , OVM or dextrane.
For optimize isomer separation , should we know configuration chemical groups (Phenyl, -NH2, =0, -OH , -COOH, others) in isomeric centres.
PS
This may be a trade secret - then we will "pass" as they say in patience (in a card game) :(
)
Before worrying about selectivity, you need to get reasonable retention. If your compounds are all eluting at the end of the gradient, then you should try a shorter-chain column. Possibly a C8 or even a C4 or cyano column. *Then* you can start looking at selectivity if necessary.

That said, be sure that you are accounting for the gradient delay time when estimating the %B at elution.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Why do you use a complicated gradient starting at 0%, when your analytes elute with 100% B?
Without further knowledge I would do a 98% B run (2 % buffer just for ionisation) and the run gradient90%B to 98% B in 25 min. 80% B to 98% B in 25 min and so on.

Alex
Maybe I am missing something but why are you derivatizing if your molecules are hydrophobic and you end up doing mass spectrometry? Are your compounds really isomers/isobars (before derivatization), meaning the same MW?
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry