Hi all,

I have no experience on ion exchange chromatography. I have the method, the column and other things. But I have no idea how to clean the system after running, in case of one night stop and several days.

Thanks for any help.
In principle, not really any different from ion-pair:
- follow the column manufacturer's recommendation for a storage buffer.
- it takes about 10 column volumes of buffer to get complete physical wash-out of a column.
- allow longer equilibration when going from a strongly bound ion (e.g., sulfate) to a weakly bound ion (e.g., chloride)
- try not to leave chloride-containing buffers stagnant in the system overnight.
- watch out for microbial growth -- filter your buffers daily and prepare fresh frequently.
- never run while holding pointed objects like scissors in your hand ;) .
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you very much !
3 posts Page 1 of 1

Who is online

In total there are 20 users online :: 2 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry