hello to all of u,
presently i am working on ETFAA and analyse it on capillary column with inj. temp. of 290°c , but when i studied with varying inj. temp. i got that 2-peaks which was showing on 290c is continuous with lowering in inj. temp. reduced and at 150°c it was reduced from 1% to 0.1%.

I dosen't understand which injector temp is suitable b' some impurities are not eluted with low inj. temp.
also the column bleeding is continuously increasing.
i just want t know is this stationary phase is suitable for this compound or not.
injector degradation
From your description you want to analyze a compound that will degrade in a hot inlet, but need the higher temperature to see some contaminants? If this is the case, you need to consider using a temperature programmable inlet or an on-column injection. These will allow the compounds to go onto the column of interest at a low temperature and elute from the column without (or with minimal) thermal degradation and then allow high boilers to also elute from the column.
3 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry