I am not a GC expert, so could use some assistance. We analyze our reactions using a DB1701 GC column on a Shimadzu GC 2010 Plus instrument with FID. Up until the issue started, we used a 17 method that has a temperature ramp of 20C/min and then keep a hold for about 5 min. The compound of interest for the method eluted as 2 peaks at 12.5 and 13.5 minutes.
This week, the peaks for the same compound shifted almost 10 minutes and began eluting at 22 and 23 minutes at 270. There has not been a change to the method at all in this time period, and nothing out of the norm has been run on the column. The solvent we are using is dioxane and methanol mixture. The other weird thing is that none of the other peaks on the chromatogram have shifted at all, just this compound. Changed the septum, insert, and put in a different DB1701 and it didnt help. Any ideas?
There is a discrepancy here between the temperature programme rate and elution temperatures (unless there is a very long isothermal hold at the beginning). At 20C/min the peaks used to elute around 260C hotter than the start temp. Even with a quite low start temp of 60C that is an elution temp of 320C, which is high for the column, and higher than the 270C (?) given as the elution temp with 10 min extra retention.
A near doubling of retention makes me suspicious that a setting for the type of carrier gas (H2 vs He vs N2) has been changed on the GC - this is usualy not set in the method and so might escape notice). Why the other peaks did not shift ?? - if they were a long homologous series with variable peak sizes from run to run it might be difficult to see whether thy had moved or not.