width vs slope?

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

5 posts Page 1 of 1
hello all,
I would like to know more about how to deal with this tow parametres..slope and width.
I am using LC SOLUTION (shimatzu HPLC).
Slope tells the integrator how much baseline rise to expect. A high slope value means that peak starts will be marked later than a low slope value, all other things being equal. For CR501s, usually start w/ 60 or 70 for slope and see if that works. Width tell sthe integrator how wide it should expect peaks to be. Too wide and it will end peaks very late. Too small and it may try to make 5 or 6 peaks out of what should be 2 peaks. Use the approximate width, in seconds, of a small peak of interest, if not the smallest.
Thanks,
DR
Image
thank you very much, I greatly appreciate your kind help,
but what do you mean "all auther things being equal"?
run conditions, gradient rise in baseline, peak size & resolution... that sort of thing
Thanks,
DR
Image
run conditions, gradient rise in baseline, peak size & resolution... that sort of thing
ok DR, I was woried about peak size..thak you very mmuch for all.
5 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry