juddc » Thu Mar 31, 2011 3:25 pm
Danko's likely correct as to the difference between the two, however we can't possibly have any hope of telling you how to "fix it" unless you give more information. From what lfew details there are, I'd guess you have a reversed phase column and your molecule is basic. If you want your analyte to be retained to a greater extent, lower the % acetonirile at the beginning of your gradient or perhaps use methanol instead, lower the column temperature, use a more retentive column (C18 versus C8, higher carbon loading, etc...), use more TFA or a higher MW ion-pairing agent (heptafluorobutyric acid) or a different ion pairing agent (alkyl sulfonates - be careful with these, they can be tricky).
To really help you we need a few details!
- Column type (and dimensions)
- Mobile phase and gradient details
- Analyte (class may suffice) with pKa data if available
- Column temperature
- Injection volume
- What you've tried before
Cheers!