Dear all,

I am having problem with auto injector (Shimadzu), the calibration results are given below:

injection volume Average Area (5 injections)
10µL 6049443
20µL 11902452
30µL 18135529
40µL 21045159
50µL 24891309


from the above results its very clear that the from 40µ onward the response is very low hence the coefficient corelation is failing.

Can any 1 suggest me the remedy....

Regards
Mohammad
Did you known the linearity range of the coffeine solution concentraion?
You can reduce the concetration of the test solution and try again.
Also, what size loop is installed on the injector. Because of "laminar flow" effects, it is not unusual to loose some sample when your injection volume gets above about half of the loop volume.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 42 users online :: 0 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry