For as long as I have worked at my current position, we have run all our RP LCs with H2O/ACN mobile phase, both with and without 0.1% formic acid additive (pH 2-3). With a bit of equilibration time between acid and no acid methods, there have been no problems. Now we are looking at having to run a method with high pH, using straight ACN, but 10mM NH4CO3/H2O (pH 9-10) as the aqueous mobile phase component.

I've never worked with this kind of solvent system before. I assume it is safest to get a dedicated column for this method. Is it possible to run both a high pH method and a low pH method on the same system? The system is an Agilent 1100 with a quat pump, so I would plan to have the channels as follows:
A - H2O + 0.1% formic acid
B - ACN + 0.1% formic acid
C - 100% ACN
D - 10mM NH4CO3/H2O (this is where 100% H2O currently resides)

Is this safe? I am concerned mostly about mobile phase carryover when switching columns, as there is always some solvent in the system. Right now a dedicated LC for this different solvent system is not an option, unfortunately. Any help, advice, correction, practical anecdote, personal experience, etc. is appreciated!

Thanks,
Kate
Presumably you saw this?:
viewtopic.php?f=1&t=16155

What is NH4CO3?
I would definitely want to dedicate columns, one to acidic and one to basic. You will also want to flush the system through with 100% water as part of a switching procedure. So if you want to run both methods in a sample list then having 100% water on your quat pump would be helpful, if you are doing this you would need to have a column switching valve too though and I assume you don't have one as you didn't mention it. In which case you could do a good manual flush through after having taken the column off before putting the other column on.

I don't see there being any inherent problems, just be careful.
Check out my blog chronicling the life of a demo Gilson Prep LC system.
http://plc2020.wordpress.com/
Presumably you saw this?:
I did. My question was more geared towards the switching between pH, but I will keep that in mind.
What is NH4CO3?
That should have read "NH4HCO3" -- Ammonium bicarbonate. Even with proof-reading things get missed...
I would definitely want to dedicate columns, one to acidic and one to basic. You will also want to flush the system through with 100% water as part of a switching procedure. So if you want to run both methods in a sample list then having 100% water on your quat pump would be helpful, if you are doing this you would need to have a column switching valve too though and I assume you don't have one as you didn't mention it. In which case you could do a good manual flush through after having taken the column off before putting the other column on.
We do have a column switching valve, with space for two columns (Agilent G1316A Thermostatted Column Compartment). I guess I was just trying to simplify things for myself, but seeing as how I am going to have switch the second column out to run this method anyway, I'll use that time to also flush the system as suggested.

Thanks for confirming what I suppose I already knew. Any other advice for a novice is always appreciated!

--Kate
4 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], John Guajardo and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry