Determination of Glutamic Acid by HPLC
Posted: Mon Mar 28, 2011 1:22 pm
by gavin.fernandes
Hi All,
I am trying to detect the presence of glutamic acid in my protein hydrolysate.
Has anyone tried this before using an HPLC method?
If so, can you please recommend: -
Column:
Eluent:
Flow rate:
Injection volume:
Detection (nm):
What retention time was obtained?
Regards
Gavin
Re: Determination of Glutamic Acid by HPLC
Posted: Mon Mar 28, 2011 1:39 pm
by Gerhard Kratz
Column: Sumichiral OA-5000 Column 150 x 4.6 mm
Part No:CH0-1859 from Phenomenex
Dimensions:150 x 4,6 mm ID
Elution Type:Isocratic
Elution A:2 mM Copper (II) sulfate in water
Elution B:Methanol
Gradient Profile:
Step No. Time (min) Pct A Pct B
1 0 85 15
Flow Rate:1 mL/min
Col. Temp: ambient Detection: UV-Vis Abs. - Variable Wave. @ 254 nm (22°C)
SIELC has also some nice applications, and other manufactureres as well. Good luck
Posted: Thu Mar 31, 2011 8:20 am
by gavin.fernandes
Is it not possible to use a C18 column and only one mobile phase?
Gavin
Posted: Thu Mar 31, 2011 12:38 pm
by Gerhard Kratz
Application Note SI-01214 Effective Determination of Glutamic Acid ... - [ Diese Seite übersetzen ]Dateiformat: PDF/Adobe Acrobat - Schnellansicht
HPLC is often used in assays for the determination of glutamic acid. In this example, an isocratic HPLC method was employed to analyze ...
www.chem.agilent.com/Library/applications/SI-01214.pdf
The chromatography conditions might be a good starting point. I would recommend to do SPE before going to HPLC, to remove all proteins in your sample. Good luck.
Posted: Sat May 14, 2011 9:54 pm
by Kostas Petritis
Check out the first publication at the following post:
http://chromforum.org/viewtopic.php?f=1&t=6509
Posted: Mon May 23, 2011 5:36 am
by Vlad Orlovsky
Posted: Sat May 28, 2011 5:57 am
by XL