Advertisement

ICS 3000

Basic questions from students; resources for projects and reports.

3 posts Page 1 of 1
Hii
I am useing ICS 3000 Dionex,
what is the best way to do calibration for bromate
There are a couple of norms around e.g. USEPA Method 300.1.
From that method or others you get information how to prepare stock solutions;

"7.3.2 Bromate (BrO3-) 1000 mg/L: Dissolve 0.1180 g of sodium bromate
(NaBrO3, CASRN 7789-38-0) in reagent water and dilute to 100 mL in a
volumetric flask."

These then can be diluted to the concentration required. Better would be to use certified standard solutions if available.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Thank you very much