Negative ion tuning problems on a Sciex API 3200
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
While performing our mandated monthly tune I'm finding that one of the instruments will show only the peak for the 49amu mass, showing nothing for the higher masses. We're using the 3x10-4 PPG Neg solution and positives look fine and pass specs. No amount of fiddling with instrument parameters seems to help. Maybe a problem with Q2 not fragmenting?
-
- Posts: 218
- Joined: Thu Mar 31, 2005 5:33 pm
Check your DAC for each mass and compare them to last tuning. But don't make any change before you know how (call the service hotline ) and feel comfortable about it. Hope this helps.
-
- Posts: 4
- Joined: Sat Feb 05, 2011 11:24 pm
thanks! The problem wound up being the offset for each mass. By moving them way up and down we were able to find the peaks and then fine tune them into acceptable specs.
3 posts
Page 1 of 1
Who is online
In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 40 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 40 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science