Agilent 1100 OQ/PV

Posted: Wed Mar 16, 2011 6:24 pm
by ksharp
Hi everyone,

I am running the Instrument OQ/PV on my Agilent 1100. I keep getting an error on my gradient composition and noise/temperature stability tests. All test result values are -3.40282e+0.38, and the status is always "error".

Any theories as to why I get this strange value for all test results?

Thanks

Re: Agilent 1100 OQ/PV

Posted: Wed Mar 16, 2011 6:54 pm
by ksharp
Oh, figured something out. I have a Sedex 75 installed on my system via HP 35900C. So I entered Configuration Editor, removed it, and re-executed the noise tests with no errors.