Agilent 1100 OQ/PV

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi everyone,

I am running the Instrument OQ/PV on my Agilent 1100. I keep getting an error on my gradient composition and noise/temperature stability tests. All test result values are -3.40282e+0.38, and the status is always "error".

Any theories as to why I get this strange value for all test results?

Thanks
I love deadlines. I like the whooshing sound they make as they fly by.
Oh, figured something out. I have a Sedex 75 installed on my system via HP 35900C. So I entered Configuration Editor, removed it, and re-executed the noise tests with no errors.