leak in mass spec

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi,

Sorry to ask again, but i am still dealing with this machine (HP 5971/GC 5890). Now, i can obtain more or less a proper signal (water and air < 5% of peak 69), but as soon as i increase the temperature of the oven I observe an increase in the air and water levels. If I decrease the temperature the levels are OK again, after a while. Could i have a pore in the column which is widening with the increase of temperature?
Any help is welcome

Thanks

Jose

Tighten the fittings while the oven is hotter, there can be expansion/contraction issues. Focus on the MSD interface, the 1/2 nut holding the gold seal and its washer in place, the weldment retainer nut, and septum nut. The issue is in the GC, not the MSD, you've already isolated that issue.
I will try replacing ferrules and setting another column... we'll see what can i obtain.

Many thanks!

Maybe gas trap or gas cylinder is contaminate.When oven temp increase, water trap on column head is liberate.Bypass gas trap...what is the result???
Good luck!
4 posts Page 1 of 1

Who is online

In total there are 104 users online :: 1 registered, 0 hidden and 103 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 103 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry