Thanks Rod,
It is a Varian 450-GC. I have a MS 5A and a porapak Q column in there. what i am doing now is bypass the MS column and just run the sample through the porapak. The GC injects samples from a SS tube connect to the sample inlet of the sample valve.
I always operate the oven at 80C with TCD 200C and filament temp 220C. I set the range to be 0.05 to get the most sensitive out of that possible.
I have notice this when i was calibrating the GC with pure CO2, but then i thought it might be some minor N2 content in the CO2 cylinder.
I have been using this GC to get CH4/CO2 seprated perfectly with TCD and FID. Now when I try to get N2/CO2 separated, the 2nd peak is really my problem.
I have also tried to ignore the 2nd peak and just take the first one as N2. Then I run pure He through to get the 'base' area. so when i have real sample injected, i will just substract this 'base' area. It seems to give me reasonable results. But I just want to sort out the 2nd peak issue.
Best regards,
George
Reconditioning the column should be of no help whatsoever with a split peak. The packing is spherical porous plastic beads (a dense styrofoam might be a closer description). Unless you have contaminated the beads with (to use a scientific term ) 'junk', there is little use in cooking out the column, it only accelerates the normal shrinkage of the beads which can produce voids and even more split or ever more increasing broader peaks.
IF, it is a column cause, then there is a void, a partial blockage, or an excessive dead space in the column or system.
But as usual, sample content, temperatures, amount injected and how it is injected all have bearing on the correct deduction of the problem.
Your question is like asking. What causes my car to 'stumble' when I drive it under 25 mph. There are a number of causes but without more information, all you will get is guess.
best wishes,
Rod