LARGE peak at beginning of GCMS run

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

8 posts Page 1 of 1
I have a large peak at the beginning of my GCMS chromatograms. I am running 5973/6890 with a Tekmar PT and Stratum. I thought it was a solvent peak, but it's there when I just run water through - it does magnify when MeOH is added to the sample. I am trying to run 524.2, but this peak is covering my first 6 compounds.
I'm running a DB-VRX (30x0.25)
Initial temp of 40 (4.00min hold), increase 10deg/min to 180, increase 15deg/min to 225 (hold 1.0min)

Does anybody have any suggestions on how to get rid of this peak? I'm fine with a solvent peak, but this one is overlapping my analytes of interest.

Thanks!
First step - get an MS spectrum of the large peak and identify what chemical it is.

Peter
Peter Apps
My money's on CO2. I always see it in purged water samples. The spectrum is mostly 44 correct?
Nothing you can do except increase your solvent delay time ( if you can), or learn to live with it.
I do not currently see it now using a 0.18 column. I would recommend your switching to a 20M X 0.18 X 1.0 column for your application. You can reduce the amount of water getting to the source ( belive me it causes problems).
Also ignore what Tekmar says and use a Vocarb 3000 trap ( you can get them directly from Supelco now).
Also try to limit the amount of MeOH you purge with carbon based traps, they don't trap as much MeOH but do a very good job of foccusing what they trap!
If it is CO2, set your MS to scan from 45 m/z on up until just before acetone. Then at least you won't have to look at it.
It is mostly CO2! Also, I am using a Vocarb 3000. I pulled out the GC uinlet last night - it was VERY dirty, didn't know if that would make a difference. I will try running some blanks on it today and see what they look like. I will also look into getting a smaller diameter column.
Won't I lose some of the lighter compounds if I set my MS scan to start at 45m/z?

THANKS!
The lowest ion I have before acetone is 45 for a qualifier for diethylether.
Acetone needs to go lower, 43 m/z, so you have to change your scan before you get there. I include Freon 113 so I scan from 45 to 200 for the first segment.
Try starting temp of 35C. I use that and my gases come out after the CO2.
Hi
I writing from Mongolia.
I have same problem with my chromatogramma.
Pls write me how to resolved your problem.
My english is too bad sorry kkk.
and send me your data acquisition method.
TNX :cyclops:
8 posts Page 1 of 1

Who is online

In total there are 84 users online :: 3 registered, 0 hidden and 81 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider], Google [Bot], Majestic-12 [Bot] and 81 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry