wan » Wed Mar 09, 2011 4:05 pm
Tom: Thank you for that info!
Petrus: Thanks for telling me that! However, I see those peaks even when I am injecting only cyanuric acid or only melamine. I did test their solubility in the mobile phase before I inject and they are soluble at the concentration I am using. I filtered my samples, thinking there are particulates not observable to the naked eye, but I am still seeing the peaks before the solvent front.
I am using the conditions stated in LIB4421, rather than LIB4422. But I have to reduce the flow rate to 0.2 mL/min, because there are leakage problems. When I increase the flow rate up to 0.6 mL/min, the pressure is less than 400 bar (232 bar), but the capillary tubing kept getting "pushed out" of the PEEK fingertight fittings, resulting in leakage. I am using 0.17 cm diameter stainless steel tubings with PEEK fingertight fittings. Is that suitable?
I understand I have to adjust the gradient according to the column dimensions and flow rate. I am trying to troubleshoot this so I can increase the flow rate and decrease analysis time.
In addition, I see that the re-equilibration time in LIB4421 is about 2.8 min, considering column volume. I back calculate the time I should set for re-equilibration based on the adjusted flow rate, but realised I cannot get a relatively flat baseline. I had to increase the timing by almost double to obtain a satisfactory baseline for the next run. This resulted in a longer analysis time. Am I missing anything?
I hope to gather more data, so I can email with the chromatograms attached. Hopefully, I will get more instrumental time to troubleshoot.