I’m not experienced in work with size exclution column, so I’d like to ask for your help. What solution (will just water be sufficient) should I wash / stabilize SEC column before first use. Is it enough to wash the column after use (mobile phase – buffer) with water only or should I use another solution?
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
The exact conditioning/initiation of the column migth be somewhat product dependent (you didn't mention the type and the make) but generally these columns are stored with aqueous mixture containing some (20 - 30 %) ethanol for conservation purposes. So you'll need to flush the storage liquid out prior to equilibration and actual chromatography. It will be a big suprise to me if it's not OK to do the flushing with pure water (some 3 - 4 column volumes) and then equilibrate with your working buffer.
After use you can store it in water if it's a short term starage (a couple of days) and water/ethanol mixture for long term storage.
Best Regards
After use you can store it in water if it's a short term starage (a couple of days) and water/ethanol mixture for long term storage.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
As danko said, it depends a lot on the type of column and on the type of samples (proteins? polymers?). It's always dangerous to reason from the general to the specific; I would suggest that you ask the column vendor about their recommendations for that specific column.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3
- Joined: Mon Feb 28, 2011 7:06 am
It's TSKgel G4000SW column, in the notice to users there is only information about washing solution after analysis and storage solution.
Thank you for your help
Best Regards
Thank you for your help
Best Regards
-
- Posts: 442
- Joined: Mon Jun 13, 2005 7:03 pm
This is from the column brochure for a Beckman Ultraspherogel SEC column which I believe is similar to the TSK-SW columns.
"Thoroughly flush the column with 20 column-volumes of fresh, HPLC-grade water and run 10 column-volumes of water or buffer (pH 3-7) containing 0.05% sodium azide to inhibit bacterial growth. Do not use halide salts in the storage buffer."
Remember that sodium azide is toxic and use due precautions
"Thoroughly flush the column with 20 column-volumes of fresh, HPLC-grade water and run 10 column-volumes of water or buffer (pH 3-7) containing 0.05% sodium azide to inhibit bacterial growth. Do not use halide salts in the storage buffer."
Remember that sodium azide is toxic and use due precautions
5 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science