-
- Posts: 45
- Joined: Thu Mar 03, 2005 11:11 pm
Mobile phase A: 0.1% TFA in water
Mobile phase B: 0.1% TFA in acetonitrile
Flow: 1.5 mL/min
Gradient: 5% B to 40%B in 5 minutes, then to 90%B at 15 minute.
We observed a sudden drop in baseline (signal goes from 150 mAU to about 20 mAU in 0.2 minute, on Agilent HPLC system). After that, the baseline drifted down slowly while the %B was increasing (from 20 mAU to negative 30 mAU in 7 minutes). This happened on different columns (same type), and on different HPLC systems. It shows up in both blank and sample injections. And it doesn't matter if we use the DAD (reference on or off) or VWD detector. When the gradient is changed, the time when this drop happens will shift but always around the same %B. We have an impurity peak that sits right before that “cliffâ€
