I am having an issue with our dionex system, and I was wondering if I could get some advice?
I am seeing a fair drift in peaks, which are eluting off much earlier than usual, also peaks that usually come off near to each other, are now forming a split peak. The method has not been changed at all.
I have tested the following
• Re-made all mobile phases and injector solvents, to rule out any uncertantity.
• Back Flushed Column Procedure.
• Replaced Column With a brand new column (improved chromatography but the problem still remains)
• Checked pressure, baseline is smooth and pressure stable.
I run with 80& H20 and 20% ACN at 1.5 ml/min
Injector solvent is 90%H20 and 10% ACN.
I am now seeing my final peak has eluted at 7.075 mins, compared to how I usually observe of 13.92 mins.
I would really appreciate some advise to how to overcome the problem. Could it be the mixing chamber? Flow cell?
Kind Regards
James