Hi. I am using a Varian Saturn 2000 GC/ion tap MS system to analyse for PAHs. I am able to get all peaks except the last 4 peaks of an EPA 17 component test mix. Theseae are the benzo(b and K)fluoranthene, benzo(a) pyrene, and the indeno upto dibenzo(ghi)perylene compounds.

I am using a a VF5 column (30m x 0.25um ft x 0.25um duiameter. In addition there peaks often seem to be shifting. AM not sure if it is a method setup or extraction setup.

Can anyone advise why I am unable to elute the last 4 peaks(m/z 252,252,276,278).

Thanks

R
What is the upper temperature of your temperature ramp and what is the temperature of your transfer line to the MS?
The instrument settings are: trap = 150 deg celc, t/fer line = 180 and manifold 170 deg celc.

I am certain these temperatures can be increased, but it was recommended during a routine maintenance by our supplier to keep the instrument settings as above.

Thanks
Apologies, typing error, manifold is set at 70 deg celcius...
If the end of your chromatographic column passes throug the transfer line, high boilers may be trapped in a cold sport at 180 degrees. I am not familiar with your instrument, so I can not make specific reccomendations.
If the end of your chromatographic column passes throug the transfer line, high boilers may be trapped in a cold sport at 180 degrees. I am not familiar with your instrument, so I can not make specific reccomendations.
That would be my thought as well. We normally go to 280 deg. on the transfer line.
I am beginning some work on PAHs right now as well. Just curious, what type of ionization source are you using on your MS?

We were having a similar problem not getting the chrysenes, while having no issue with our lower MW PAHs.
It ended up being our extraction procedure. The method we got from the literature did not have us collecting enough from an Alumina and then a Silica Column. We were leaving the higher MW compounds behind
what you temperature program and you column. I go to 300 degrC for db-5MS UI
difficult to say definitively as gc paramters unkown but t/f line T seems way too low for heavy PAH.
9 posts Page 1 of 1

Who is online

In total there are 54 users online :: 0 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 54 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry