Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi all,
I had recently come accross a intresting problem which i could not able to solve yet.
I was analysing a protein compound(inclusion bodies). It is a basic analytical method using C4 column with water and Acetonitril TFA. When i injected the standard i got the response which i usually get but, for the sample i got only half the response as compared to my regular response. As a matter of fact, i had just injected in to another HPLC wherein, i got correct sample response as expected. I do not find any difference between two HPLC's(both are Agilent 1100). My sample is slightly viscous as compared to the standard so, the i first thing i checked was injector draw speed & eject speed but both the HPLC's were set same(100µl/min).

Can any one comment on this?
One system to another? Could be something other than autosampler. First question is "how is the precision of replicates" ?

Tubing dimensions, tubing dead volumes, detector cell differences, VWD v. DAD, binary v. quad pump, going through thermostatted compartment tubing, one having a column-switching valve, lots of things could be the reason.
Along with the thoughts presented by CPG - needle offset? Maybe one is at 1 mm and the other at 0, and you happened to bottom out in the 0-offset vial? Or perhaps you have a leaking metering seal, or had a partially plugged injection seat, or maybe a failing rotor seal? Like the man said, there can be a lot of factors. If you can consistently take a standard from one instrument to the other and obtain similar results, but taking a sample (I mean the exact same vial in both cases!) between the two results in a 50% difference, then this is a possibly more involved problem. Hopefully you re-prepped and re-injected your sample on both instruments?
Time flies like an arrow. Fruit flies like a banana.
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry