In order to prevent overloading of the IC column due to the high chloride concentration I use onguard Ag cartridge. I am afraid that Ag traces will damage the IC column. That is why I need to add onguard cartridge in H+or Na+ form. Which of them is preferable?
H+ is more easily displaced than is Na+
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
That is clear. But H+ catridge will reduce pH.
Before injection, can you put the sample solution into a syringe and pass through a silver cartridge to remove the chloride BEFORE injection? We've used Grace/Alltech Maxi-Clean Ag to do that, worked good.

I admit that I never did assay for trace Ag+ in the filtrate though...
That is clear. But H+ catridge will reduce pH.
Ok, so you have to decide the extent to which that is a concern. My guess is that unless you are injecting a *huge* sample, it should not be a problem.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts Page 1 of 1

Who is online

In total there are 34 users online :: 2 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry