Sandra7 » Thu Feb 10, 2011 5:44 am
well, I will give you some advice to choose a better wavelength.
The wavelength you chose must have a good response to both component.
If we just have one component (usually when we want to quantitate the component), we choose the maximum absorption wave length which usually have the best response to your component.
If we have many component to detect, we should choose a wavelength which can have a response to all of the components, so you can scan every of your standard from 200nm to 400 nm, to detect in which wavelength rank they all can have a good response.
Also you can have a DAD chromatomap if you have a DAD/PDA dector, which maybe helpful for you to choose a better wavelength.
Remember, the wavelenthe you choose must have a relative good response to all of your components. The wavelenth you choose if very important when you want to develop a HPLC method.
Best wishes,
Yours Norman
Yes, I will search for the best wavelength for all 6 standard compounds because Professor wants a better analysis of the standard mixture that shows better peak for benzene We (my lab partner and I) were supposed to do that before trying to run analysis for quantification for our unknowns (we each have separate unknown sample). She was using the instruments for quantification runs as soon as she got her unknowns identified and so we got behind on improving separation of the standard mix. The availability of the instruments outside class hour is limited.
BTW, I did analysis of benzene standard alone with the same method for acetophenone. It elutes later than acetophenone. So, polarity was the primarily effect.