The grob mix injection program I used was;
isothermal at the PTV inlet 260oC with a glassfrit inside the liner.
split 1:100 from 0 to 2 min, flow 1 mL/min.
Column oven; initial 40oC (hold 2min), 180oC (10oC/min), 300oC (60oC/min, hold 2 min), total 20 min.
I looked at the split DCM peak once again with the MS spectras; (sorry my bad) - the first tailing peak at 1.4 min is butane

(which I injected not ages ago, but it should have been swept away...??). The second tailing peak at 1.5 min looks like some kind of halogenated degradates (maybe from DCM combined with other residues)...
In the system there is an EFC, however, this device was (maybe faulty) installed from the vendor as outlet to 'atmospheric' conditions and not 'vacuum'. I have tried to alter the setting, however, then the EFC cannot calibrate the flow

Why I have turned it back to 'atmospheric'. But I guess, I agree with Rod, the unretained peak time should be something like 1.4-1.5 min in this system.
I'm running a condensation test this evening to check if there is a contamination from gas lines and injector.
hope you have some suggestions, thanks
Martin