I'm busy analyasing Mebeverine HCl by HPLC. The standard and sample are dissolved in water. During the analyses we have discovered that the area of the mebeverine peak decreases over time. We have tried using amber glassware/vials, but the area still decreases over time. The BP 2011 has a uv method for mebeverine but we would like to use a stability indicating HPLC method. The BP method ueses 0.1M HCl as the solvent for the UV. We will try this solvent to see if mebeverine is more stable in this solvent. Has anyone experienced similar problems with mebeverine?

Thanks in advance.
Mike
Hi

If you goggle stability mebeverine, you will fina a number of articles.

This one indicate that silanols may be an issue:
http://www.springerlink.com/content/ft31x44630luh73k/

Have not worked with the substance myself but seems to be a bunch of articles around for stability analysis by HPLC.

regards
chris
Izaak Kolthoff: “Theory guides, experiment decides.”
2 posts Page 1 of 1

Who is online

In total there are 20 users online :: 3 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry