- Posts: 12
- Joined: Tue Jan 25, 2011 5:55 pm
I have two peaks from placebo that interfere with my main peak. They are not from diluent and the system is clean. The overlay of placebo and API shows these peaks are right underneath the API. My assay method is isocretic, ACN/MeOH/H2O=800:190:10, diluent is Ethanol:H2O=90:10, flow rate :1.5mg/ml, column temp: 22degree. 3.5x250mm C18 column. My current retention time is 11min.
I can't increase H2O ratios in mobile phase because my diluent would be stronger and I assume that would mess up my peak shape. (I don't have room to play with diluent due to the nature of the sample). I prefer not to change column and hope to seperate these interference peaks from main peak without too many trail and errors. So:
1. Which parameters shall I try first?
2. If change ACN and MeOH ratio what ratio do you suggest to start with? I don't know huch much 1% organic change will affect retention time? Does anyone have an idea?
3. If change temperature, what suggested temp shall I go for?
Thanks a lot.