Tetrahydrofuran x RP Columns

Posted: Thu Jan 20, 2011 2:54 pm
by carlos.teixeira
Hi Folks,

The method below has produced constant overpressure after 8h to 14h of pump flow (analysis). The pressure starts with 170 kgf/cm2 but always the chromatograph stop before the end of the analysis,since the max. pressure is 400 kgf/cm2.

1) Do you have any ideia about the cause?

2) Is it common any THF generate column pore swelling?

Chromatographic method
Mobile phase composition: MeOH = 90%, THF = 5%, Water = ยจ5%
Flow rate: 1.5 mL/min
Temperature: 25 degree Celsius
Column: C-18 Zorbax 250mm 4.6mm 5um

Thanks,

Re: Tetrahydrofuran x RP Columns

Posted: Thu Jan 20, 2011 6:51 pm
by Consumer Products Guy
We've used THF for years with RP-18 columns, and not had any such issues. If you haven't already, run the sequence but WITHOUT actual injections for your 8 to 14 hour timeframe, and see if the pressure still goes up like you stated. And post what you find.

Re: Tetrahydrofuran x RP Columns

Posted: Mon Jan 31, 2011 1:15 pm
by carlos.teixeira
Sorry for my delay... we just changed THF for ACN and problem was resolved.

Wonderful elutions for everybody!!