Tetrahydrofuran x RP Columns

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi Folks,

The method below has produced constant overpressure after 8h to 14h of pump flow (analysis). The pressure starts with 170 kgf/cm2 but always the chromatograph stop before the end of the analysis,since the max. pressure is 400 kgf/cm2.

1) Do you have any ideia about the cause?

2) Is it common any THF generate column pore swelling?

Chromatographic method
Mobile phase composition: MeOH = 90%, THF = 5%, Water = ¨5%
Flow rate: 1.5 mL/min
Temperature: 25 degree Celsius
Column: C-18 Zorbax 250mm 4.6mm 5um

Thanks,
We've used THF for years with RP-18 columns, and not had any such issues. If you haven't already, run the sequence but WITHOUT actual injections for your 8 to 14 hour timeframe, and see if the pressure still goes up like you stated. And post what you find.
Sorry for my delay... we just changed THF for ACN and problem was resolved.

Wonderful elutions for everybody!!
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 4 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Baidu [Spider], Semrush [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry