Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
A question about tailing: This is for assay testing. Normal concentration and injection volume (5uL) gives 2.119 tailing. When injection amount changed by injection volume (uL) it gives inj.vol/tailing values 2/1.556, 3/1.772, 5/2.119, 7/2.316, 10/2.617. Clearly at normal condition tailing is too high, when injection amount increased tailing also increases. Is this not acceptable (for a good reliable assay method)? What are the risks? Does this variation suggest column overloading?
Ranjith
Overloading is one possibility. The other is the "strong solvent effect" if the diluent is stronger than the mobile phase (ideally,the diluent should be the same or weaker).

What column dimensions?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 3 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry