kerri » Tue Jan 18, 2011 1:04 am
Ok, sorry for the misnomer on DMHA (just not familiar with it). Whoops I made a fool of myself
Secondary and tertiary amines are proton-abstracting modifiers and, like fredklink explained, they will cause signal suppression in positive mode. You might have better luck using plain old 0.1% ammonium hydroxide to help your signal instead of DMHA... but a buffer is best as you have complete control of the pH.
I copied the LC/MS conditions of the paper I mentioned previously and pasted them below. These conditions (or modification thereof) are definitely worth a shot.
LC–MS Instrumentation and Conditions
An Agilent (Palo Alto, USA) 1100 MSD single quadrupole mass spectrometer equipped with an ESI source operating in the positive ion mode was used as a detector...[clip] The selected ion monitoring (SIM) mode was used for quantification with m/z of 508 for ATP and 428 for ADP. The ESI–MS operating variables used in this study were as follows: capillary voltage, 3.0 kV; source temperature, 300 °C; drying gas flow, 10 L/min; fragment voltage, 100 V; and nebulizer pressure, 30 psi.
The LC separation was performed using an Agilent Extend C-18 column (3.1 x 100 mm, 3 um). The mobile phase consisted of 5 mM ammonium acetate (pH 7.5) and methanol (60:40, v/v), and it was delivered at a flow rate of 0.3 mL/min; the column temperature was set at 20 °C.
One question, do you make your ATP solution fresh?