Possible detector problems on 5890 ecd

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
After about a month the 2nd channel starts to lose sensitivity and resolution, so I change the column, which I just did last week. I ran once and the peaks were shaped beautifully, but when I tried a second run the 2nd channel went bad again. I ran a bunch of rinses thinking there might have been some debris, and then ran a low level standard. It had gotten worse. When I take the cover off the top of the detector it looks like there is a copper sleeve on the first detector that isn't on the second one, and the post coming out of the top of the second one is loose. Anyone have any ideas?
Image
Image
Sorry about the huge photos. No photo editing program at work.
JReed,

The second one appears to be missing the insulator that isolates the collector anode from the body. Therefore, you can either be leaking through this fitting causing noise or you can be wiggling and grounding through this fitting (causing noise.)

Best regards,

AICMM
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry