We need to validate an analytical RPC method that is used for detection of % oxidation of the main analyte in terms of peak area % relative to the main peak. We have a specified limit of 8% for our product which when exceeded the product is considered non conforming. In other words, we need the method to be able to quantify the % (in terms of % peak area) when it is around 8%, but below that, accuracy is not of concern, as long as it didn't reach 8%.
I have tried to check ICH Q2 guidelines but I can't decide under which category the method should fall: whether it is quantitative or limit test for impurities? Both require different validation tests, but in my case I feel I need a combination of both:
1- Specificity (oxidation product baseline separated form the main analyte, anything else to validate here?)
2- LOD, but also I would need to prove accuracy at the 8% level, somehow!
3- Linearity, range, precision: do I really need to perform these in my case?
What also increases the challenge is that we have 3 oxidation products, 1 arises first, followed by the other 2 with further oxidation, so to isolate the impurity "standard" to use for spiking, should I work with a mixture of all 3?? I personally think I should only use the 1 that 1st arises for both simplicity and also to be more realistic as this is the one we will usually see, the other 2 appear only after aggresive oxidation with H2o2.
I would really appreciate your opinions and suggestions on how to approach this validation, and what should be required to be proved from such a method.
I would also appreciate literature suggestions for such specific type of methods.