Just now I come across with one journal that managed to separate all tocols in RP-HPLC within 18 minutes using caratenoid column C30 250x4.6mm, 3um.
More detail in
Influence of Growth Temperature on the Amounts of Tocopherols, Tocotrienols, and ç-Oryzanol in Brown Rice
J. Agric. Food Chem., Vol. 55, No. 18, 2007 pg 7561
100% solvent B (acetonitrile/tetrahydrofuran/water 70:25:10, v/v/v) between 0 and 25 min
linear gradient to 100% solvent A (acetonitrile/tetrahydrofuran, 50:50, v/v) from 25 to 35 min
100% solvent A from 35 to 45 min
ending with 100% solvent B from 45 to 60 min
25 °C ,1.0 mL min-1
However, the peak separation is not good with low resolution specially beta and gamma peaks.
What can I do to increase the resolution of these peaks to get better separation.
I am scare to upload the chromatogram as it is from journal.
And what the disadvantages of C30 column over C18 column
Hope will get good reply from you all.