System suitability

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Which criteria do you use to define the system suitability test for intermediates in a synthetic route for an API ?.....

Should be the same as the product ?..

Thanks

Same kind of specs (retention windows for compounds of interest, resolution between critical pairs, plus the usual repeatability, pressure, etc.) but appropriate to the compound for which you are analyzing.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Tom,
do you record pressure only for potential troubleshooting? Also what in particular would you note when using a gradient - the initial pressure prior to an injection, maximum over an injection?

Yes, primarily for troubleshooting. Probably the initial pressure would be the easiest/most relevant as an indicator of column problems.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry