In autotune i m getting peak width adjust error.
In manual tune no effect through width gain and width offsets in scan mode i m getting very wide and noisy peaks. I did ion optics cleaning during cleaning My ion optics assembly octopole rods soldering was damaged. Octopole rods free from Soldering and now my RF is not stable and its fluctuating.
whats the reason and solution of wide and noisy peaks.
How can I Solder these rods?
Is it possible?
Any other required info will available.
Please suggest

Sorry for my english.
Thanks
My answerer is the obvious. You need to contact Agilent. You instrument is literally broken. :cry:
Regards;
David

O. David Sparkman
Consultant-At-Large
My answerer is the obvious. You need to contact Agilent. You instrument is literally broken. :cry:
thanks for your reply David

i have changed octopole. And Now i m gettng proper RF and MSD Passing all diagnostic test.
But Still autotune showing "Peak width adjust Error" and wide and noisy peaks in manual tune scan mode.

Thanks and Regards
Prishi
I don't know if this will help or not, but sometimes if an instrument's tune file has got a very long way away from where it ought to be, the autotune can't find its way back. If it's possible to reset the factory defaults, it can be helpful to do so, and then run autotune (save your tunefile before resetting).

I believe reset to factory defaults is somewhere in the file menu of the tune page.
I don't know if this will help or not, but sometimes if an instrument's tune file has got a very long way away from where it ought to be, the autotune can't find its way back. If it's possible to reset the factory defaults, it can be helpful to do so, and then run autotune (save your tunefile before resetting).

I believe reset to factory defaults is somewhere in the file menu of the tune page.
thanks for ur suggestion
i Have reset factory defaults but no progress.
I m gaining mid masses like 622 on 608 and 922 at 898 with a good absorbance but low mass 118 showing very low abundance and high masses having noise there with a high abundance.
If u need picture of this than i should provide u but tell me how?

i m Suspicious about Desolvation assembly. It may be a cause for wide and noisy peaks ? please suggest me how can i check desolvation assembly.?

Thanks a lot again
5 posts Page 1 of 1

Who is online

In total there are 188 users online :: 0 registered, 0 hidden and 188 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 188 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry