Protein injection in C18 column
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
I am working of detecting protein ligand complex. Can I inject the sample directly in my C18, 4.6x150mm,3.5micron column, assuming that ligand will separate from protein as the organic solvent will kick in and i can detect it.Will the protein destroy my column....can somebody help me i am very new to this protein thing.
-
- Posts: 345
- Joined: Sun Dec 06, 2009 6:16 pm
Proteins are routinely analyzed with RPLC so its unlikely your protein will damage your particular C18 column. Be sure to elute the protein from your column to avoid plugging. I have run myglobin using an Onyx monolithic C18 column (50x2.0mm) and a gradient from 5-95% ACN with 0.1% formic acid without problem. The heme and the apomyglobin were well resolved.
A wide pore material (300A) may be needed depending on the size of the protein your analyzing. If the protein is too large for standard HPLC media (~100A) it will simply elute very early or unretained.
Hope this helps
Carl
A wide pore material (300A) may be needed depending on the size of the protein your analyzing. If the protein is too large for standard HPLC media (~100A) it will simply elute very early or unretained.
Hope this helps
Carl
A. Carl Sanchez
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
It also depends on the nature of the protein. If the protein is very hydrophobic (has many hydrophobic residues) a C18 column may be too hydrophobic and you may have difficulty getting it off the column. From what I understand, C8 columns are more frequently used with whole proteins. Also ion exchange, and size exclusion are also frequently used with proteins though more so for preparative applications.
I've worked with membrane binding proteins that were so hydrophobic that putting them in just about any column was like the protein bermuda triangle (It disappears and is never seen again).
I've worked with membrane binding proteins that were so hydrophobic that putting them in just about any column was like the protein bermuda triangle (It disappears and is never seen again).
-
- Posts: 1
- Joined: Wed Dec 29, 2010 9:19 am
Dear,
Try ZICHILIC COLUMN FROM MERCK WHICH IS VERY SUITABLE FOR PROTEIN SEPARATION THIS COLUMN IS FROM MERCK
Try ZICHILIC COLUMN FROM MERCK WHICH IS VERY SUITABLE FOR PROTEIN SEPARATION THIS COLUMN IS FROM MERCK
4 posts
Page 1 of 1
Who is online
In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 24 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science