Contamination to MS from Mol Sieve PLOT 5A column?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I am about to install an Agilent Mol Sieve 5A into a 5972 MS.
30 M, 0.32 mm, 0.25 um coating

Are there any precautions to consider?
I am concerned about particles from the PLOT column
contaminating the 5972 MS.

Do you suggest a downstream trap?
Would a "column quick connect" into the MS be a good idea?

Welcome your suggestions.

Ed
I would recommend you install after your column a piece of thick film methyl silicone narrow bore capillary column, 1 micron 0,25mm ID a few meters of length would do. This acts as a sticky trap for any particles broken off from the sieve.

best wishes,

Rod
+1 on the particle trap. I installed a PLOT column in our MS against the advice of our service rep. I added a particle trap and didn't have a lot of problems. I should add that I valved the GC so I could select either a DB1 column or the PLOT column. I always left the DB1 column "in service" with the PLOT purged to vent whenever the GC/MS wasn't used. This minimized the chance that a stray particle would enter the MS
3 posts Page 1 of 1

Who is online

In total there are 49 users online :: 0 registered, 0 hidden and 49 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 49 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry