H2 gas analysis with TCD, Argon carrier gas

Discussions about GC and other "gas phase" separation techniques.

8 posts Page 1 of 1
I have submitted gas samples for GC analysis for O2, N2, CO2, CO, C1-C4 hydrocarbons and H2S to two different labs. The results of analysis between the labs (using direct sample injection from canister to instrument from duplicate samples), both using a GC FID/TCD, do not agree. One lab reports H2 gas 2-3 orders of magnitude greater than the other.

The analytical procedure for the one lab, reporting [H2] of 1700 ppm is as follows:

Sequence of analysis: CO2, O2, N2, C1-C4 were run then the sulphurs and finally the H2.

- The samples were not diluted when the CO2, O2, N2, C1-C4 were run. The canisters were pressurized just before the sulphurs were analyzed.

- GC Model: 7890A GC FID/TCD

- Oven temperature: Run starts at 40 and goes to 200 at varying ramps. H2 comes out between 140 and 200

- Single inlet

- Argon Carrier Gas

- 1 mL Sample loop volume

- Backflushing time: 0.5min with carrier gas

- Column specs: Supelco 60/80 Carboxen 1000 15’X1/8” SS (2.1mm ID)

- 2X dilution with ultra high purity nitrogen

The other lab reports [H2] of 810,000 ppm. I am waiting to get details of analytical procedure for this lab.

Can you suggest possible gas interferences, or analytical errors that might cause this 2-3 order of magnitude discrepancy in Hydrogen gas?

Thank you!
1 - mis-identification of valve upset peak as a hydrogen peak

2 - calibration mix error

3 - human error

You realize that asking this question without GC hardware and parameters and chromatograms displayed is a big waste of time ? IMHO.

Where's the BEEF ?

best wishes,

Rod
810,000 ppmv is 81%, what are the concentrations for N2, O2? The total should be close to 100%.
Do you have any knowledge of the sample composition that would allow you to rule out one of the results? H2 may elute very near a valve actuation depending on the GC configuration (as chrom1 points out).
I am very surprised to see that you think that hydrogen elutes between 140°C and 200°C with a starting temperature of 40°C. Hydrogen is hardly retained by any column that I know of, and I would expect it to elute VERY early in the chromatogram. This looks like misinterpretation as mentioned by chromatographer1.

Gasman
While this column does retain hydrogen to a small degree, more than almost any other packing I know, with a 6m 2mm ID column at 120°C with helium carrier at 30cc/min (45 psi) the retention time of hydrogen is about 1.62 min.

best wishes,

Rod
Gasman

I think you were misled by the numbers w/o a label. I think they meant 140 to 200 seconds.

Rod
Post the cromatograph of both TCD detector. I assume that there is a data missinterpreted.
8 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry