- Posts: 25
- Joined: Mon Dec 06, 2004 11:32 am
I am having problems measuring low concentrations of trihalomethanes (<100ug/L) by GC/FID. This is because I cannot get a stable baseline for some reason, and there are background peaks occuring near to where the THMs peak. I am using a headspace extraction method, so there could be contaminants in the air, but I have tried different types of air (lab, fume cupboard, outside, zero-grade air from cylinder) and they all give similar readings.
The use of a smaller syringe (100uL as opposed to 500uL) results in a much more stable baseline. Does this rule out column bleed as a reason for the peaks when using a 500uL syringe? Unfortunately, I need to use the larger syringe size for THM analysis.
If anyone has any suggestions what could be causing the peaks I would be very grateful.
I am using a capilliary column, 100C isothermal, 200C injector, 350C detector.
Many thanks