Hi,
Thanks you for your replies.
I don't have a lot of experience with HPLC but if the injection loop volume was greater wouldn't the retention times differ?
I am using a brownlee aquapore column and guard column with C4 packing. I am using buffers:
A: 10% MeCN / 90% Milli Q water / 0.1% TFA
B: 90% MeCN / 10% Milli Q water / 0.1% TFA
flow rate: 0.8 ml/min
injection volume: 100 uL.
My gradient is:
Time (minutes) %B
0 0
4 0
6 28
11 48
14 55
15 58
18 60
18.01 0
If the injection loop volume is greater wouldn't it take longer for the more concentrated MeCN to get to the column and hence longer for the proteins to elute?
I have just ran some more samples and the compounds are eluting about 0.5-0.6 minutes later, with the peek tubing injection loop, than they used to with the stainless steel injection loop. For example one major compound was eluting at ~8.9 minutes (with stainless steel injection loop) and now it is eluting at ~9.5 minutes. We have made up several lots of buffers and we keep getting the same results.
This seems too big a difference. Even if it was due to the different size injection loop, the retention time should only differ by: 13 uL / 800 uL/min = 0.016 min.
Does anyone have any ideas?
Cheers,
Laudrup