Hi Amy, the BIOQ triple quad MS should already be an MS/MS instrument or does MS/MS not work on this machine? Since this is an old Micromass MS it should already have a photomultiplier detector so from that point sensitivity should be ok

.
MS/MS will definitely increase sensitivity in a complex sample matrice since it will decrease noise. It should improve your LOD probably at least to a factor of 10.
The more up-to-date triplequad machines from Waters/Micromass all have the Z-Spray ionsource which also increases sensitivity but also robustness in dirty samples.
If you don't get your machine running to the performance you need you could look for another lab in your university which have a newer triple quad MS which you could possibly use for a limited time period. All the triples which are about 5 years old or younger should be from a factor of 50 to more than 100 more sensitive than your old BioQ and developping a MRM method for stress hormons should not take much time. Once you have got the method the analysis of the samples should really go fast...