Acquity PDA Baseline Signal Bottoming Out (negative)

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi there,

I've had this issue before, and cannot remember how to fix it. Basically the detector signal goes negative and "flat-lines", and then doesn't seem to be a way to "balance" or rectify. The signal is just staying flat at -0.0653 AU!

Can anyone offer any suggestions.

PS. Not running anything unusual: big standard water, AcN +0.0%% TFA in both lines, gradients - or trying to ;-)

All wavelengths?
Any error messages?
If you reboot the system, does the problem persist?

My first guess would be an electronic problem. The instrument should run a self-test at power on.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Maybe you started up your detector with an air bubble in the cell.
Flush the system with methanol, leave a flow through the detector and cycle power of your PDA.

Ace
3 posts Page 1 of 1

Who is online

In total there are 279 users online :: 1 registered, 0 hidden and 278 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 278 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry