What's the appropriate amount of sample to load on in column chromatography? Is there a rule of thumb?

I have a column that is 4.5 cm x 27 cm

silica/THF

Thanks !

Is there a rule of thumb?
For "analytical-scale" separations (non-overloaded), the rule of thumb for a long time was that you could to up to about 100 micrograms on a 250 x 4.6 mm column. That would imply about 10 milligrams on your column (10x the diameter implies 100x the volume).

The catch is that prep separations are almost always run under "overload" conditions. How much you can overload on depends on the selectivity between your product and the closest interfering impurity.

I usually suggest a loading study on an analytical scale column (e.g., 250 x 4.6 mm), going up in geometric progression (e.g., 30, 100, 300, 1,000, . . .) to establish the maximum load for *your* sample. Then scale that up to the prep column in proportion to the column volume (in this case, 100x).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry