Guys,

Can you please help me. I'm not able to pass FID detector temperature accuracy test. My specification says 250C + or - 10C where as I can reach only 235C.

What do i need to do If I want to pass this test?


Thanks
Sakthi

have you lost insulation? Do you have the thermometer probe at the proper location? Have you considered re calibrating your temperature settings to compensate for the lower temperatures produced?

If you don't have any mistakes, then adjust your hardware to get the proper temperatures. Just because the dial may read 250, it doesn't mean it means 250, it means what you get.

Rod

You should have no problem getting higher than 300 C, something is wrong if you can't get past 235.
have you lost insulation? Do you have the thermometer probe at the proper location? Have you considered re calibrating your temperature settings to compensate for the lower temperatures produced?

If you don't have any mistakes, then adjust your hardware to get the proper temperatures. Just because the dial may read 250, it doesn't mean it means 250, it means what you get.

Rod
hi,

"then adjust your hardware to get the proper temperatures" - how to adjust the hardware? Is there is any manual available? I'm using Agilent GC.

Thanks for your reply.

If you set the detector temperature by computer, and you set it to 250C but only get 235C by true and actual measurement with a thermometer, at what setting would you set the detector temperature to achieve a true temperature of 250C ? or 235C? or 275C? or 150C ?

Spend a hour or two and determine the actual temperatures achieved at different settings and document it.

Use the document as a calibration of your instrument.

Unless you wish to perform your own service, ie. replace controller boards, or determine why the sensors are shutting off the heater at the wrong temperature.

What else are you going to do, other than buying a new GC ?

If you get 235C when you type in 250C, and you don't want 235C , then don't type in 250C. Type something else instead.

You have to determine the proper setting to get the temperature you want.

Or call in the service department and pay for the work they will do.

Good luck,

Rod
5 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry