Baseline decreasing with temperature program
Posted: Tue Oct 26, 2010 1:13 pm
				
				Hi there 
I need a help please.
I am working with GC-FID Focus GC ThermoFinnigan, a GC column DB5-MS from Varian, with the following gradient temperature program:
0-2 min 40 ºC
3ºC/min until 175 ºC
5ºC/min until 250 ºC
250ºC for 15 minutes
Injector 250ºC
Detector 260ºC
Split flow 1/40
He 1,5 mL/min
N2 as make-up gas
After 7 minutes of running the program the baseline drops.
I have checked possible leaks, the flow rate of the gases are stable, the setpum is new and I think the column conditions are appropriated.
Can you help me with this problem?
Can I quantify the peaks even with this drift?
Thank you in advance
Best regards
Rui