Hi there

I need a help please.

I am working with GC-FID Focus GC ThermoFinnigan, a GC column DB5-MS from Varian, with the following gradient temperature program:

0-2 min 40 ºC
3ºC/min until 175 ºC
5ºC/min until 250 ºC
250ºC for 15 minutes

Injector 250ºC
Detector 260ºC
Split flow 1/40
He 1,5 mL/min
N2 as make-up gas

After 7 minutes of running the program the baseline drops.
I have checked possible leaks, the flow rate of the gases are stable, the setpum is new and I think the column conditions are appropriated.

Can you help me with this problem?
Can I quantify the peaks even with this drift?

Thank you in advance

Best regards

Rui
Rui Gonçalves

is it a gradual decline? Lots of runs will show some baseline drift. You'll just have to look at the integration and see if the drift looks detrimental to the results.

One possibility is that you are running column compensation. If the baseline profile changes, you could have a baseline drifting lower. If that's the case, you may need to rerun the column compensation.

Maybe a good cleaning of the column?

I still want to say it's a leak. You're absolutely positive there are none? Typically the baseline goes up with temperature though... so it may not be.
GC-TCD/NPD (Agilent 7890)
GC-MS (Agilent 6890)
GC-TCD/uECD (HP 5890) - "Ole Miss"
GC-TCD (Carle)
GC-TCD/FID (SRI)
IC - (Dionex ICS-3000 + AS1/ERG)

Thank you all for your help
is it a gradual decline? Lots of runs will show some baseline drift. You'll just have to look at the integration and see if the drift looks detrimental to the results.
We only made a few runs, the decline apear all the time, don´t increase with the number of runs.

I will se again if I have any leak.
Could be also the fact that the column don´t reach the detector?
Rui Gonçalves
5 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry