Advertisement

PFPD vs. FPD

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi all,

I know the principle of FPD detector. But recently see the term PFPD meaning Pulsed Flame Photometric Detector. So what is the 'pulsed' here ?

Thanks

thory,

The PFPD keeps the fuel/air ratio below that needed to maintain a constant flame. Then you leave the ignitor on all the time. This results in a build up of flammable gases followed by ignition and propagation but not sustained. Pop. The pop results in excited state species for sulfur or phosphorous (or others) that then emit, similar to an FPD. The emission of the sulfur is delayed relative to the emission of the carbon species. If you take advantage of the delay and only measure the signal in the time window of the sulfur emission you can increase your sensitivity and selectivity.

The main disadvantage is you get about 3-5 pops per second so you are rate limited on the data collection. This limits the ability of the PFPD to be used with narrow peaks or high speed GC.

Those are the basics....

Amirav has lots of articles in Analytical Chemistry for your review.

Best regards,

AICMM
2 posts Page 1 of 1

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry