I’ve only been doing HPLC (the reverse phase style) for a few weeks now, so please bear with me and talk to me like a child. I’ve been having some difficulties this week with respect to not getting “peaksâ€
-
- Posts: 303
- Joined: Fri Sep 03, 2004 8:06 pm
What is the "new" detector make and model? Perhaps you only have the UV lamp turned on, and not the VIS lamp, so as not to be able to go to the longer wavelengths? Check for leaks!
Good luck!
Good luck!
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
One always looks at the point where problems started to get at the cause, in your case it seems obvious: the methanol is too strong a solvent for the sampe, it "washes in" the analytes, or at least part of them, giving these broad peaks. There are a lot of mentions of sample solvent incompatibilities (with the mobile phase) in this forum.
3 posts
Page 1 of 1
Who is online
In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science